How to choose a laser for Raman Spectroscopy
In specific spectroscopy techniques like the Raman or LIBS (Laser-Induced Breakdown Spectroscopy) ones, the choice of the color and of the optical specifications of the excitation laser can be very important.
The Raman spectrum of a material or a gas sample represents the energy shift of its vibration modes in relation to the laser excitation and is usually expressed in wavenumbers (cm-1 unit).
In Raman spectroscopy, the effective radiated power varies as the excitation frequency to the fourth (due to the electric dipole nature of light), therefore proportionally to λ0-4 of the wavelength λ0 of the Raman laser. Thus it may be interesting to use shorter wavelengths to increase the signal to be analyzed, typically with 514 nm or 532 nm green Raman lasers.
Visible Raman lasers are especially used for inorganic material experiments and for Surface Enhanced Raman Scattering (SERS).
However, the excitement of some samples may be accompanied by a phenomenon of fluorescence that can interfere, even exceed the Raman signal. This problem can generally be solved by considering a lower laser energy, and therefore a higher wavelength, typically with 785 nm or 830 nm near-infrared Raman lasers. Recent developments can extend this consideration to the infrared range with 980 nm or 1064 nm infrared Raman lasers.
Red and NIR Raman lasers are especially beneficial for the study of bio-molecules where good fluorescence suppression is needed.
The possibility of multiplexing measurements at different wavelengths – 532 nm, 785 nm and 1064 nm – could also be a considerable advantage for the analysis of the sample properties.
Figure 1. Typical Raman spectra of Diamond (left) and Cyclohexane (right) samples with 785 nm laser excitation (spectral density versus Raman shift): diamond spectrum presents a single-mode peak on a mean fluorescence signal; cyclohexane spectrum is a multi-mode signal.
In addition to the wavelength and the power of the laser, several optical specifications are important to consider while choosing your Raman laser:
- the stability in power is an important criterion in the choice of the exciting laser, especially for repeatable and long duration measurements: good Raman lasers have a power stability better than ± 1 % peak-to-peak in a 8 hour period.
- the laser linewidth and the stability in wavelength (“laser drift”) of the Raman laser are also important for the quality of the spectrum, especially for high spectral resolution reconstruction: good Raman lasers have less than 0.5 % rms of optical noise.
- the lifetime of the Raman laser is another critical parameter to check: 10,000 hours of expected operation is a minimum to have.
- for specific applications, the polarization extinction ratio of the Raman laser must be controlled: a ratio greater than 100:1 is generally expected.
Figure 2. Typical high-resolution spectrum of a low-end multimode high power Raman laser: the linewidth provided by the vendor corresponds to many modes and may be quite unstable.
7 Comments
Very nice paper !
But, no think is said about the linewidth of the laser … any indication?
0.1 nm is good? 0.5 nm?
Dear Amayas,
Thank you very much for your comment.
You are right, the linewidth can be an important specification to consider for a Raman laser, and must be linked with the spectral resolution you need on the Raman spectrum.
Good single-mode Raman laser diodes can typically provide very narrow linewidths, less than 100 MHz FWHM (i.e. 0.0002 nm at 785 nm) with a wavelength stability better than 0.010 nm, but are relatively low in power (50 to 100 mW); combined with a high-resolution spectrometer, they can be very valuable for high-coherence Raman spectroscopy to differentiate very small Raman shifts.
Multi-mode Raman laser diodes provide higher power (up to 600 mW), but their linewidth can be as wide as 1 nm; so this results in all Raman peaks being 1 nm wide, regardless of the resolving power of the spectrometer.
A good tradeoff can be a single-mode Raman laser with 0.1 nm linewidth and ~ 300 mW intermediate power.
Hoping that it will be useful for you,
Best regards,
Fabrice
Again …. a wonderful answer !
Thank you Fabrice …
Here are some very practical questions!
So, can you help to decide whiche one is the best for Raman spectroscopy among these lasers?
1- BeamQ laser
http://www.beamq.com/green-line-generator-laser-532nm-200mw-p-530.html
2- Lasever laser
http://www.lasever.com/plus/view.php?aid=11
3- eBay Laser
http://www.ebay.com/itm/200mW-532nm-Green-Laser-Dot-Module-TTL-Analog-0-30KHZ-TEC-Cooling-85-265V-/111693314992?hash=item1a0170bbb0
Knowing that they are all 200mW adjustable with variable power supply
they are all 0.1 nm of linewidth.
the major difference is the satability, the BeamQ one has 3;5% over 8hr times, while the others are 5% per 4 hr
So, as for me, i will decide for the BeamQ even if it is more expensive with 300$
knowing that they are all chineese lasers, do you think that paying 300$ is worth it ?
an other alternative solution, more relaiable and more expensive of course, is the thorlab solution (diode laser https://www.thorlabs.de/newgrouppage9.cfm?objectgroup_id=5597&pn=DJ532-40) plus the power supply, TEC module … all will cost 1800€ .. much much more expensive.
Please, Fabrice,
tell me what you think
Amayas
thank you again for the help
Dear Amayas,
Yes, 300-600 dollars is currently the price of chinese low-end multi-mode lasers ; they may be generally sufficient to spectrometric studies with 0.5 nm resolution (20 cm-1 Raman shifts at 532 nm).
The difference may come from the higher power and the stability of the BeamQ. Better quality will require a price > 2 k$ .
If you use fibered or bulk-optics setups, be careful that all these lasers are generally free-beam and not necessarily collimated (see beam divergence and quality), so you will need collimating optics (with basically 40-60% coupling efficiency).
If you want more information, please feel free to send me a direct email to : fabrice.thomas@resolutionspectra.com
Best regards,
Fabrice
Dear Fabrice !
THanks a thousand for these precious information.
I’ll be back to you by email
All the best for the MyLaserSpectrum community
Amayas
I mentor high school science fair projects and a student is considering a green laser pointer as a Raman spectrometer source. I’ve always thought that the optical noise of these $10 pointers make them unsuitable, however I see a 2012 reference to paper presented to the Optical Society of America using a laser pointer for an Raman explosives detector. Is using a cheap laser pointer wishful thinking? or does the student have a good change of detecting Raman spectrum?
Hello Allen,
It is true that new green laser pointers can now achieve very high power, more than 1 Watt. So maybe it can be potentially used for low-precision Raman detection, but it will require a very good optical system for beam focalisation and Raman signal observation (with a centered laser rejection filter).
Be also very careful that this kind of very-high-power green lasers are very dangerous for eyes and bodies (use of specialised protection glasses, do not point towards persons or objects like inflammable material, planes,…) and are sometimes reserved to military applications. For more information, please refer to the Code of Federal Regulations (especially 21 CFR 1040-1041) of the CDRH in the US.
Best regards,
Fabrice